This may have been asked already, but does anyone chart vein size and compressiblity, and if so, how do you measure and how do you decide if the vein is big enough for dual, triple. Then how do you chart it?
We started doing this more than a year ago. Currently, we are charting the vein size and its compressibility without the tourniquet, before the insertion.
Based on our ultrasound machine, we either use the size of the vein in its natural state (without tourniquet) in approximate cm. or on the french size that displays on our SiteRite. The SiteRite has a panel on the right side with the comparable fr. size that is compared with the display that we see. We chart this info in the documentation of the PICC. We do not insert multiple lumen PICCs in patients with small veins that are the same size of the PICC. We try to determine a vein at least twice the size or 3 times the size of the PICC.
Anecdotally, I believe that this practice has decreased the number of DVTs that occur. We have a study approved to prove that. We hope to continue to decrease the occurrence of DVTs associated with PICC insertion.
Some times it is tough to argue with the nurse or doctor about the number of lumens, but we have stuck with our parameters. They want a triple and we don't see a vein that will tolerate that. We put in what the vein can handle.
we have been charting vein size and vessel compressibility for about 2 years now as well. We generally follow the same rule as outlined in the earlier response (no more than half of the vessel should be consumed by the PICC line). We take a firm stand on vessel size especially with our 6FR triple lumen catheters as well. As a general rule, we follow these guidelines
Cath/FR sizeMin Vessel Size
3FR = 1.0 mm (.1cm).20cm - .30cm
4FR = 1.35mm (.135cm).27cm - .41cm
5FR = 1.67mm (.167cm).33cm - .50cm
6FR = 2.0mm (.2cm).40cm - .60cm
7FR = 2.3mm (.23cm).46cm - .69cm
8FR = 2.7mm (.27cm).54cm - .81cm
9FR = 3.0mm (.30cm).60cm - .90cm
Catheter size of picc should be 1/3 to ½ the diameter of the vessel.
I was wondering where this chart came from. Is it from a manufacturer? My hospital really wants evidence to back up what we do. Thanks for providing it!
No this didn't come from a manufacturer, I created the scale myself based on the idea of the catheter not exceeding more than 1/2 the vessel size - so for a 5FR catheter which is 1.67mm (or .167cm) in diameter, you would want your vessel size to be at least .33cm (twice the size of the diamter of the catheter), preferably .50cm (3 times the size of the catheter) or even larger. I did get the diameters sizes from an educator with Navalyst (formerly boston scientific)
Thanks for the chart. I would also like to know your resources? We are starting to follow this practice as weel. We get pressure, especially from ICU, do place multiple lumen PICC's that aren't appropriate for the patients vessel sizes. Any supporting documentation will help. I appreciate it. Robin
We started documenting vein size in natural state. This is french size (based on siterite) but also have an ilook. For the ilook we have calipers that can measure diameter but I didn't know how to correlate this to french size. The chart looks helpful, what is the source?
French size is the outer diameter of the catheter measured in mm then multiplied by 3. So an OD of 2 mm would be a 6 French catheter. This will help you compare this to the measured vein size. My question is are you actually measuring the outer vein diameter or the vein lumen diameter? Lynn
I am a clinical specialist for Teleflex and initially began researching catheter vessel ratio about 6 years ago during my clinical practice at a large teaching facility. I saw a relationship between the size of the cathter and the occurrence or upper extremity thrombosis (UEDVT) when our team investigated a newly designed PICC which was introduced to the marketplace. Placement of any catheter in vessels create some degree of stasis. It is exaggerated with the placement of larger lumen catheters. I played around with some numbers over the course of many months and for some reason the "1/3" rule seemed to make sense to me. 1/2 of the vessel seemed too much and 1/4 seem unrealistic. It was then that my interest in thrombosis surfaced and I have been entertained by it's occurrence, prevention and management ever since.
A colleague of mine, Dr. Tom Nifong also has a huge interest in thrombosis; he manages a hematology thrombosis lab in Hershey PA. He and I have presented all around the country on this very topic. With the dramatic increase in the placement of PICC's in hospitalized patients, Tom began investigating the very question.....where is the data that supports the 1/3 cathter vessel ratio rule? He has done studies with flow models determining how much the blood flow in the vessel is decreased with the insertion of various sized catheters. His initial data supported the 1/3 rule, that being that the catheter not take up more than 1/3 of the vessel. As you can see this is only a two dimensional measurement, as that is easiest for many ultrasounds to determine, but actually the measurements in his caluclations is determined by flow characteristics in the entire lumen of the vessel.
An easy way to remember what size the vessel should be as it related to a particular catheter is as follows:
3F PICC needs a 3mm vessel (or .3cm)
4F PICC needs a 4mm vessel (or .4cm)
5F PICC needs a 5mm vessel (or .5cm)
6F PICC needs a 6mm vessel (or .6cm)
This formula works for catheters that have true French sizes from tip to hub--meaning a catheter that is designed with a non-tapered configuration. If your PICC is reverse tapered, you will need to alter the size of the vessel component on the calculation, as a catheter that is for example a 7F at the point that enters the vein should have a 7 mm vessel and so forth.
I might add, that someone mentioned in the thread about 1/2 of the vessel size.....Tom's calculations showed dramatic decreases in blood flow when the catheter consumed more than 1/3 of the vessel.
Lastly, I think that Tom's data will be published in the near future, and thus, the information we all have been asking for will be available. Also, to let you know, a 6 session thrombosis lecture series will soon be launched, suitable for nurses or physicians. Tools to assist the clinican with appropriate devcie selection and thrombosis tracking thru out catheter dwell will be available as well. I might also add that Dr. Scott Trerotola will be presenting a lecture relative to catheters and UEDVT at AACN conference in May in Washington DC, if any of you are CCRN's.
First of all, I enjoy reading all the comments shared and value the various feedback on postings. I have a question as I am somewhat confused after reviewing the chart presentation on cath size and minimal vessel size. It is helpful and a good tool however, the OD of the 4FrSL picc I am using (per package description) is 1.45mm and for a 5FrDL 1.80mm. The chart indicates that the 4Fr is 1.35mm (recent change to 1.33) and the 5fr is 1.67mm. I assume there are some differences depending on product/manufacturer and maybe the # of lumens? I would think that when the OD is known as obtained from the package, it's always best to perform the calculation from that OD for vessel size? If anyone has any information about the variance in cath diameters in relationship to the chart and my Picc OD measurement or any other useful information on the cath to vessel size debate, I would appreciate some feedback.
The outer diameter of all catheters are measured by a straight line from the outside wall through the center of the lumen to the outside wall of the opposite side. There could be slight differences between brands due to the thickness of the catheter wall. Silicone catheters have a much thicker wall than a polyurethane catheter. The configuration of the lumens could be another reason for this variation. Lumens can be even split by adding a center wall to form to D shaped lumens. They could also be elliptical or oval shaped. You would also want to go by the measurements provided by the manufacturer for the specific brand you are using. This measurement is taken in millimeters. The French size is converted from this mm measurement. MM measurement X 3 = French size. So a 6 Fr catheter would have a measurement of 2 mm OD. For comparison to the internal diameter of the vein wall, there is one published article assessing mathematical formulas for the blood flow surrounding the catheter. I am not aware of any clinical studies on outcomes about the practice yet. So should the catheter consume only a third of the vein lumen or is a half acceptable? This is an unresolved issue based on good solid evidence. Of course it makes intuitive sense that the smaller catheter consuming less of the veins ID would be better but still the science is missing on this issue. Lynn
Try contacting your PICC product rep; they might be able to locate it.
Nina Elledge, RN, MBA, CRNI
[email protected]
We started doing this more than a year ago. Currently, we are charting the vein size and its compressibility without the tourniquet, before the insertion.
Based on our ultrasound machine, we either use the size of the vein in its natural state (without tourniquet) in approximate cm. or on the french size that displays on our SiteRite. The SiteRite has a panel on the right side with the comparable fr. size that is compared with the display that we see. We chart this info in the documentation of the PICC. We do not insert multiple lumen PICCs in patients with small veins that are the same size of the PICC. We try to determine a vein at least twice the size or 3 times the size of the PICC.
Anecdotally, I believe that this practice has decreased the number of DVTs that occur. We have a study approved to prove that. We hope to continue to decrease the occurrence of DVTs associated with PICC insertion.
Some times it is tough to argue with the nurse or doctor about the number of lumens, but we have stuck with our parameters. They want a triple and we don't see a vein that will tolerate that. We put in what the vein can handle.
Gwen Irwin
we have been charting vein size and vessel compressibility for about 2 years now as well. We generally follow the same rule as outlined in the earlier response (no more than half of the vessel should be consumed by the PICC line). We take a firm stand on vessel size especially with our 6FR triple lumen catheters as well. As a general rule, we follow these guidelines
Cath/FR size Min Vessel Size
3FR = 1.0 mm (.1cm) .20cm - .30cm
4FR = 1.35mm (.135cm) .27cm - .41cm
5FR = 1.67mm (.167cm) .33cm - .50cm
6FR = 2.0mm (.2cm) .40cm - .60cm
7FR = 2.3mm (.23cm) .46cm - .69cm
8FR = 2.7mm (.27cm) .54cm - .81cm
9FR = 3.0mm (.30cm) .60cm - .90cm
Catheter size of picc should be 1/3 to ½ the diameter of the vessel.
Hi Karen,
I was wondering where this chart came from. Is it from a manufacturer? My hospital really wants evidence to back up what we do. Thanks for providing it!
Sheila
Sheila Hale, RN, CRNI, VA-BC
Sheila,
No this didn't come from a manufacturer, I created the scale myself based on the idea of the catheter not exceeding more than 1/2 the vessel size - so for a 5FR catheter which is 1.67mm (or .167cm) in diameter, you would want your vessel size to be at least .33cm (twice the size of the diamter of the catheter), preferably .50cm (3 times the size of the catheter) or even larger. I did get the diameters sizes from an educator with Navalyst (formerly boston scientific)
Thanks for the chart Karen!
Thanks for the chart. I would also like to know your resources? We are starting to follow this practice as weel. We get pressure, especially from ICU, do place multiple lumen PICC's that aren't appropriate for the patients vessel sizes. Any supporting documentation will help. I appreciate it. Robin
We started documenting vein size in natural state. This is french size (based on siterite) but also have an ilook. For the ilook we have calipers that can measure diameter but I didn't know how to correlate this to french size. The chart looks helpful, what is the source?
French size is the outer diameter of the catheter measured in mm then multiplied by 3. So an OD of 2 mm would be a 6 French catheter. This will help you compare this to the measured vein size. My question is are you actually measuring the outer vein diameter or the vein lumen diameter? Lynn
Lynn Hadaway, M.Ed., RN, NPD-BC, CRNI
Lynn Hadaway Associates, Inc.
PO Box 10
Milner, GA 30257
Website http://www.hadawayassociates.com
Office Phone 770-358-7861
lumen diameter
Sheila,
I am a clinical specialist for Teleflex and initially began researching catheter vessel ratio about 6 years ago during my clinical practice at a large teaching facility. I saw a relationship between the size of the cathter and the occurrence or upper extremity thrombosis (UEDVT) when our team investigated a newly designed PICC which was introduced to the marketplace. Placement of any catheter in vessels create some degree of stasis. It is exaggerated with the placement of larger lumen catheters. I played around with some numbers over the course of many months and for some reason the "1/3" rule seemed to make sense to me. 1/2 of the vessel seemed too much and 1/4 seem unrealistic. It was then that my interest in thrombosis surfaced and I have been entertained by it's occurrence, prevention and management ever since.
A colleague of mine, Dr. Tom Nifong also has a huge interest in thrombosis; he manages a hematology thrombosis lab in Hershey PA. He and I have presented all around the country on this very topic. With the dramatic increase in the placement of PICC's in hospitalized patients, Tom began investigating the very question.....where is the data that supports the 1/3 cathter vessel ratio rule? He has done studies with flow models determining how much the blood flow in the vessel is decreased with the insertion of various sized catheters. His initial data supported the 1/3 rule, that being that the catheter not take up more than 1/3 of the vessel. As you can see this is only a two dimensional measurement, as that is easiest for many ultrasounds to determine, but actually the measurements in his caluclations is determined by flow characteristics in the entire lumen of the vessel.
An easy way to remember what size the vessel should be as it related to a particular catheter is as follows:
3F PICC needs a 3mm vessel (or .3cm)
4F PICC needs a 4mm vessel (or .4cm)
5F PICC needs a 5mm vessel (or .5cm)
6F PICC needs a 6mm vessel (or .6cm)
This formula works for catheters that have true French sizes from tip to hub--meaning a catheter that is designed with a non-tapered configuration. If your PICC is reverse tapered, you will need to alter the size of the vessel component on the calculation, as a catheter that is for example a 7F at the point that enters the vein should have a 7 mm vessel and so forth.
I might add, that someone mentioned in the thread about 1/2 of the vessel size.....Tom's calculations showed dramatic decreases in blood flow when the catheter consumed more than 1/3 of the vessel.
Lastly, I think that Tom's data will be published in the near future, and thus, the information we all have been asking for will be available. Also, to let you know, a 6 session thrombosis lecture series will soon be launched, suitable for nurses or physicians. Tools to assist the clinican with appropriate devcie selection and thrombosis tracking thru out catheter dwell will be available as well. I might also add that Dr. Scott Trerotola will be presenting a lecture relative to catheters and UEDVT at AACN conference in May in Washington DC, if any of you are CCRN's.
Cheryl Kelley
Cheryl Kelley RN BSN, VA-BC
First of all, I enjoy reading all the comments shared and value the various feedback on postings. I have a question as I am somewhat confused after reviewing the chart presentation on cath size and minimal vessel size. It is helpful and a good tool however, the OD of the 4FrSL picc I am using (per package description) is 1.45mm and for a 5FrDL 1.80mm. The chart indicates that the 4Fr is 1.35mm (recent change to 1.33) and the 5fr is 1.67mm. I assume there are some differences depending on product/manufacturer and maybe the # of lumens? I would think that when the OD is known as obtained from the package, it's always best to perform the calculation from that OD for vessel size? If anyone has any information about the variance in cath diameters in relationship to the chart and my Picc OD measurement or any other useful information on the cath to vessel size debate, I would appreciate some feedback.
The outer diameter of all catheters are measured by a straight line from the outside wall through the center of the lumen to the outside wall of the opposite side. There could be slight differences between brands due to the thickness of the catheter wall. Silicone catheters have a much thicker wall than a polyurethane catheter. The configuration of the lumens could be another reason for this variation. Lumens can be even split by adding a center wall to form to D shaped lumens. They could also be elliptical or oval shaped. You would also want to go by the measurements provided by the manufacturer for the specific brand you are using. This measurement is taken in millimeters. The French size is converted from this mm measurement. MM measurement X 3 = French size. So a 6 Fr catheter would have a measurement of 2 mm OD. For comparison to the internal diameter of the vein wall, there is one published article assessing mathematical formulas for the blood flow surrounding the catheter. I am not aware of any clinical studies on outcomes about the practice yet. So should the catheter consume only a third of the vein lumen or is a half acceptable? This is an unresolved issue based on good solid evidence. Of course it makes intuitive sense that the smaller catheter consuming less of the veins ID would be better but still the science is missing on this issue. Lynn
Lynn Hadaway, M.Ed., RN, NPD-BC, CRNI
Lynn Hadaway Associates, Inc.
PO Box 10
Milner, GA 30257
Website http://www.hadawayassociates.com
Office Phone 770-358-7861